MULTICENTRE STUDIES

FACTORS THAT INFLUENCE THE RESULTS OF ANTI-β2GPI ELISA


 

COORDINATORS

 

Guido Reber1, Anna Cavazzana2, Philippe de Moerloose1

1Haemostasis Unit, Geneva University Hospital, Switzerland and 2University of Padova, Italy

 

 

PROJECT

 

A previous collaborative study from the Standardisation Group of the European Antiphospholipid Forum on anti-β2GPI antibodies home-made assays has shown that differences at the antigen level could explain only in part the variability in assay results. In this study, participants were asked to compare the results of selected samples obtained with their routine β2GPI preparation with a common β2GPI and two calibrators sent to all laboratories, using their local protocol. They were also asked to fill a questionnaire describing their local protocol.

On the basis of the technical data provided by the participants for each step of the local ELISA protocol, we tested the influence on assay results of pH of the coating solution, of some microplate brands, the blocking solution, the ionic strength of the dilution buffer, as well as the effect of calcium ions and of various enzyme-conjugated signal antibodies. We tested 32 patient’s samples (25 APS, 2 SLE and 5 other autoimmune diseases) that were quantified with the Forum Calibrators, the monoclonal antibody HCAL and our local (single patient) calibrator.

The influence of pH of the coating buffer, either neutral (PBS) or alkaline (carbonate pH 9.6), on patients’ OD was tested with the Stago, the Hyphen and a home-made β2GPI. There was no difference except for the home-made β2GPI for which slightly higher values were observed at alkaline pH. The binding of each calibrator to the microplate varied depending on the brand. This was also the case for some patients. In consequence, the combination of both binding patterns (calibrator and sample) resulted in different results as a function of the microplate brand. The use of BSA or gelatine in the blocking buffer, as well as omitting the blocking step, did not change mean results when using HCAL for calibration whereas the absence of blocking resulted in higher values when using the Forum Calibrators. Sample-blanks were clearly higher when using gelatine with both Costar and Maxisorp microplates. Increasing the ionic strength affected more the binding of Forum Calibrators than that of HCAL whereas the opposite pattern occurred in the presence of calcium ions. Five enzyme-conjugated signal antibodies were studied. Their pattern of binding to Forum Calibrators was more homogenous than to HCAL. Therefore, the mean value of patients results varied widely according to the conjugate and the calibrator used.

For additional information and applications, please e-mail: guido.reber@hcuge.ch


Web contens: EWPSLE Info Desk.

 

Last updated: 17 November 2005